    Default Syncrowave SD 180 no high frequency

    Folks -
    Trying to troubleshoot my old (LA 081331) SD 180. I get no HF just a little (1/8 in spark) and thats it at the torch. When I checked the spark gaps, one of them was shorted - Probably had some grit blown into it. Readjusted the gap and now nothing at the Spark Gaps of 0.012. I checked T3 without power and get continuity at the primary side and 6K ohm on the secondary (That seems a bit high to me). The cap shows open but I can't check to see if its working. Continuity on T4 windings seems good. The foot switch (switch and 1K pot) is working as well. I can hear gas come on when i hit the foot switch and there is current at the torch (spark when i hit the electrode to the work material)

    Anyone have any ideas?

    cheers,
    -jim

  2. #2
    Join Date
    Nov 2003
    Posts
    3,150

    Question

    Quote Originally Posted by jim.2435074 View Post
    Folks -
    Trying to troubleshoot my old (LA 081331) SD 180. I get no HF just a little (1/8 in spark) and thats it at the torch. When I checked the spark gaps, one of them was shorted - Probably had some grit blown into it. Readjusted the gap and now nothing at the Spark Gaps of 0.012. I checked T3 without power and get continuity at the primary side and 6K ohm on the secondary (That seems a bit high to me). The cap shows open but I can't check to see if its working. Continuity on T4 windings seems good. The foot switch (switch and 1K pot) is working as well. I can hear gas come on when i hit the foot switch and there is current at the torch (spark when i hit the electrode to the work material)

    Anyone have any ideas?

    cheers,
    -jim
    I would pull the covers and blow it out with compressed air...

    you likely have HV leakage from dust accumulation/carbon tracking..

    Have you had the HF set to continuios with the covers off in a dark room??..

    you might be able to see the blue corona of the HV leakage paths...
    .

    *******************************************
    The more you know, The better you know, How little you know......

    “The bitterness of poor quality remains long after the sweetness of low price is forgotten”

    Buy the best tools you can afford.. Learn to use them to the best of your ability.. and take care of them...

    My Blue Stuff:
    Dynasty 350DX Tigrunner
    Dynasty 200DX
    Millermatic 350P w/25ft Alumapro & 30A
    Millermatic 200

    TONS of Non-Blue Equip, plus CNC Mill, Lathes & a Plasmacam..

  3. #3
    Join Date
    Nov 2013
    Location
    Washington DC area
    Posts
    6

    Default

    Quote Originally Posted by H80N View Post
    I would pull the covers and blow it out with compressed air...

    you likely have HV leakage from dust accumulation/carbon tracking..

    Have you had the HF set to continuios with the covers off in a dark room??..

    you might be able to see the blue corona of the HV leakage paths...
    Yeah I did blow the machine out, and did a quick look for a high voltage spark leak in a near-dark garage ... did not see or hear anything ... Will do it again and pay more attention this time.

    Does 6K ohm seem right for T3 secondary? that seems a bit high for a coil winding to me.

    cheers,
    -jim

  4. #4
    Join Date
    Nov 2013
    Location
    Washington DC area
    Posts
    6

    Default

    Quote Originally Posted by jim.2435074 View Post
    Yeah I did blow the machine out, and did a quick look for a high voltage spark leak in a near-dark garage ... did not see or hear anything ... Will do it again and pay more attention this time.

    Does 6K ohm seem right for T3 secondary? that seems a bit high for a coil winding to me.

    cheers,
    -jim
    Thanks to the friendly folks at BR Welding Supplies White Plains MD, I was able to get some data on the SD180.
    1. The 6K ohm on T3 secondary was correct.
    2. looked for an HV leak and didn't find it - But did not expect to cause
    3. No 120V on T3 primary when I hit the remote pedal (reads 0V AC) - which is kinda weird cause I can sometimes get a small spark on the spark gaps / torch

    Gonna start working my way up to the main board and look for 120V ... Anyone have any ideas?

    cheers,
    -jim

  5. #5
    Join Date
    Dec 2007
    Location
    Oklahoma
    Posts
    705

    Default

    Quote Originally Posted by jim.2435074 View Post
    Thanks to the friendly folks at BR Welding Supplies White Plains MD, I was able to get some data on the SD180.
    1. The 6K ohm on T3 secondary was correct.
    2. looked for an HV leak and didn't find it - But did not expect to cause
    3. No 120V on T3 primary when I hit the remote pedal (reads 0V AC) - which is kinda weird cause I can sometimes get a small spark on the spark gaps / torch

    Gonna start working my way up to the main board and look for 120V ... Anyone have any ideas?

    cheers,
    -jim
    I believe many Miller technicians have recommended HF points gap be at .008"
    instead of the manual spec of .012".
    Probably can find this recommendation with a search on this forum.

    Griff

  6. #6
    Join Date
    Nov 2013
    Location
    Washington DC area
    Posts
    6

    Default

    OK, This is the excuse I was looking for.

    No 120V power to T3 from the main board. A replacement board is almost as much as the resale value of the machine. Will take my time / chances and look at the main board, but in the mean time ...

    I have ordered a Dynasty 280DX from Oxygen Supply Company (aka BR welding) in White Plains, MD. More Welding Power, less current draw, light weight and lots of bells and whistles!

    cheers,
    -jim

Tags for this Thread

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •  

Warning: Non-static method PhpQuickProfiler::getMicroTime() should not be called statically in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/pqp/classes/Console.php on line 77

Warning: Non-static method PhpQuickProfiler::getMicroTime() should not be called statically in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/pqp/classes/Console.php on line 77

Warning: Non-static method PhpQuickProfiler::getMicroTime() should not be called statically in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/pqp/classes/Console.php on line 77

Warning: Non-static method PhpQuickProfiler::getMicroTime() should not be called statically in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/pqp/classes/Console.php on line 77

Warning: Non-static method PhpQuickProfiler::getMicroTime() should not be called statically in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/pqp/classes/Console.php on line 77

Warning: Non-static method PhpQuickProfiler::getMicroTime() should not be called statically in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/pqp/classes/Console.php on line 77

Warning: Non-static method PhpQuickProfiler::getMicroTime() should not be called statically in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/pqp/classes/Console.php on line 77

Warning: Function split() is deprecated in /mnt/stor3-wc1-dfw1/357822/357839/www.millerwelds.com/web/content/lib/footer.inc.php on line 82

Welding Projects

Special Offers: See the latest Miller deals and promotions.